Everything in my tank is doing good, i have high phosphorus 74 ppb (0.227 ppm), I trust the checkers.
Follow along with the video below to see how to install our site as a web app on your home screen.
Note: This feature may not be available in some browsers.
Firstly because they're still closed, and second because they don't measure phosphates in my gallery. That's something we would send to water quality to get tested, and I can't send my personal tank samples to water quality testing. Not every week or as often as I'd like to test.
I too wasn't attacking, just asking why you weren't checking it at the aquarium business you work at.The checkers are toys compared to real testing equipment. But unless a hobbyist wants to spends five figures on proper equipment, we are working with what's available.
I wasn't personally attacking you. The most important thing with the checker (aside from clean vials) is time span. You need five minutes between reagent activation and test result measurement. This is why the mixing period is two minutes, plus the built-in countdown timer.
I too was having problems with the checker, but once I bought new vials, keep them very clean and follow the five minute procedure, the results are very consistent.
The checkers are toys compared to real testing equipment. But unless a hobbyist wants to spends five figures on proper equipment, we are working with what's available.
I wasn't personally attacking you. The most important thing with the checker (aside from clean vials) is time span. You need five minutes between reagent activation and test result measurement. This is why the mixing period is two minutes, plus the built-in countdown timer.
I too was having problems with the checker, but once I bought new vials, keep them very clean and follow the five minute procedure, the results are very consistent.
How do you clean your vials? - something I always struggle with.
My HI774 always around 0.08 ppm above my ICP and test with Red Sea Pro Phosphate test. Knowing that - I can use it - just take away 0.08 ppm. But it nearly cost me my reef before I understand that. Precision can be bad in some checkers - IMO - and the repeatability is not better. The HI774 have ± 0.02 as accuracy on the paper. I always use two vials - one zero and one sample. I check that both show 0.00 and keep one as zero. Put in reagents in the sample vial and do as it should be done (2 minute shaking - 3 minutes waiting). after this - I take the zero and do the test procedure to the C2 is in the window. Change tube to the sample and a short press (reading directly without the 3 min) - note. New zeroing with the zero sample - C2 - change to sample vial - short press - read. Doing this for at least 10 times - here is my last readingsI too find the Hanna Checker to be the most annoying kit I use, but there really is no choice for low nutrient measurement.
Over time, I have managed to keep test results pretty consistent. What helped:
a) replace cuvettes immediately if they get a scratch
b) nitrile gloves
c) electrostatic wiping cloth
d) 2 vials, two minute mix time
e) extra care with reagent loading; every spec goes into the cuvette
f) store vials filled with RODI
All that said, my Hanna consistently reads .02 to .04 high relative to ICP.
This isn’t the correct way to test With a hanna checker, I’d advise reading the manual, testing correctly then seeing where you are before you do anything else..I use both flasks for the test, one with tank water for C1, and the other tank water and powder, I mix the powder for 1.5 minutes, then put flask 1 in tester (c1) then it says add c2, my mix is done and ready right away...then hold for my 3 minute timer.
i´m sorry to say - if you follow the manual - it just a lottery when you analyse low levels. The one I use have a written accuracy of ± 0.02 ppm PO4 (HI774), HI713 have ±0.04 ppm and the HI736 have ± 5ppb P (around 0.015 ppm PO4)This isn’t the correct way to test With a hanna checker, I’d advise reading the manual, testing correctly then seeing where you are before you do anything else..
Test | Expected Reading (PPM) | Actual Reading (PPM) |
1 | 0 | 0 |
2 | 0.05 | 0.006 |
3 | 0.1 | 0 |
4 | 0.25 | 0.077 |
5 | 0.5 | 0.15 |
Test | Expected Reading (PPB) | Actual Reading (PPB) |
1 | 0 | 0 |
2 | 16 | 2 |
3 | 33 | 0 |
4 | 82 | 25 |
5 | 163 | 49 |