I've always found that my Hanna salinity testers show lower results compared to the Milwaukee one, even after calibrated with the Hanna calibration solution. To clarify the issue, I prepared standards for both refractometers and conductivity meters with Randy's recipe in the lab I work at and tested them.
I mixed up 6.2g of NaCl with 182g of RODI to make a 3.29% solution for conductivity.
The 3.65% refractometer standard is made by mixing 3.65g of NaCl and 96.35g of RODI.
I then brought the solutions back home and conducted the experiments on the same day.
Here are my results:
* Both Hanna testers were calibrated with their own bag of Hanna calibration solutions right before the test, and were cleaned with RODI water and dried out for 30 minutes.
* The Milwaukee tester and handheld refractometer were both calibrated with RODI water to 0 ppt before the test, and were allowed one minute for temperature equilibrium before collecting data during each test.
Does anyone have similar experiences with these instruments? It seems that Hanna has a different definition for S=35 seawater, which does not correspond to the typical 53 mS/cm at 25°C.
I mixed up 6.2g of NaCl with 182g of RODI to make a 3.29% solution for conductivity.
The 3.65% refractometer standard is made by mixing 3.65g of NaCl and 96.35g of RODI.
I then brought the solutions back home and conducted the experiments on the same day.
Here are my results:
Milwaukee | Hand-held refractometer | Hanna A | Hanna B | |
testing standards | 35 ppt (3.65% NaCl) | between 36~37 ppt (3.65% NaCl) | 33.8 ppt (3.29% NaCl) | 33.8 ppt (3.29% NaCl) |
testing my reef tank | 35 ppt | between 36~37 ppt | 32.6 ppt | 32.6 ppt |
* Both Hanna testers were calibrated with their own bag of Hanna calibration solutions right before the test, and were cleaned with RODI water and dried out for 30 minutes.
* The Milwaukee tester and handheld refractometer were both calibrated with RODI water to 0 ppt before the test, and were allowed one minute for temperature equilibrium before collecting data during each test.
Does anyone have similar experiences with these instruments? It seems that Hanna has a different definition for S=35 seawater, which does not correspond to the typical 53 mS/cm at 25°C.