As they say, "A man with a clock knows the time. A man with 2 clocks is never sure."Wow I completely believed them to, I figured a company like Hanna instruments would know what they are talking about but upon reading what you wrote about ethanol there isn’t even enough organics in our reef tanks to change the outcome of a refractometer, so meaning if they are both calibrated correctly they should technically read the same discounting there margin of error on the devices, so that leaves me with the question of witch one is right? My Milwaukee or my Hanna salinity pen? That question is difficult to answer though.
I have tried all the salinity tools available. Hanna pen, Pin Point EC meter, Icecap salinity tester, Milwaukee Refractometer, manual refractometer, etc. , all of the ones on the market, literally all of them.
My biggest disappointment was the Hanna Pen. I was so pleased when they brought it out. I thought it's Hanna, it's going to be good. Not my unit unfortunately. I can never get it to calibrate and hold a calibration. Super disappointed in that thing.
As I mentioned above, I use the TM High Precision floating hydrometers as they are the gold standard. I also scoured Ebay to hunt down some used Thermo Scientific conductivity meters. I got the idea to use the Thermo's from Randy. Those work wonderful, IF, you can find the right probe for the model you have, in the right measurement range, that still works (used). Finding working probes used is the key as they are crazy expensive new.
I have an Orion Star 3 that I use as my main meter, and a Orion 105A+ as my cross reference. The Star 3 gives a result in about 10 seconds after power on. The 105A+ takes about a minute to settle. By comparing the 2, I can tell that they are still in calibration ... or they both drifted equally
For calibrating the Orion's, I use a TM hydrometer as it does not require calibration.
NOTE: For the TM hydrometers, be careful where you source them. The ones sold in Germany do NOT measure in Specific Gravity (S.G), but instead measure in density. The readings look similar, but are not. Bad things will happen if you mistake the one for the other ... speaking for ahhh, friend.
Dennis